From the device parameter perspective the flip angle is limited by the maximum channel output level. To have a more meaningful range for the scaling factor, we would need to agree on a maximum amplitude in Hz, i.e. a maximum flip angle? |
|
I recently implemented something similar on the python version of the Kea console which is a B1 efficiency in B1 per Volt, with the voltage being the output of the RF channel of the console. If you have that it's trivial to calculate the voltage needed for any flip angle/pulse duration (assuming linearity). This is probably the way to do it here too, even if the scaling to output voltage is not immediately obvious it doesnt matter, as long as it scales consistently. This can be expanded if the gain of the RF amplifier is known, if this is a variable that can be given we know the power going in to the coil which is relevant for SAR. |
Uh oh!
There was an error while loading. Please reload this page.
Uh oh!
There was an error while loading. Please reload this page.
The RF amplitude in pulseq is defined in Hz.
When unrolling the sequence, the amplitude in Hz must be translated to a voltage in mV.
We currently have two scaling factors for the translation.
A coarse scaling factor defined in
device_config.yamlwhich brings the amplitude close to 1 and a finer scaling factor which is adjusted to the object (usually it is in the range between 1 and 10).However, the coarse scaling value was chosen empirically and the second finer scaling factor is optimized to the object.
It would be nice to have a comprehensible value which depends on the maximum output that can be achieved with a setup, so that the maximum B1-scaling value corresponds to 1.